DRS6A RTR-93-E RSB-118 low range performance with almost new magnetron MAF1422B
exsat2003
Dear All,
I am a technichian in Italy.
www.exsat.com
We have a DRS6A transceiver in our workshop that have poor performance in range.
The magnetron was replaced recently by other company and now have 320h only.
The boat is now experiencing very low range performance.
From the RADAR status page all is ok but Magnetron voltage at 2,8V !
Checking the RTR at our workshop on a dummy load for X band RADAR up to 25Kw
and spilling out the DIRect RF from a wave guide coupler with -20dB and other attenuators to a total of -80dB
the reading on our spectrum analizer is -38dBm, so -38dBm + 80dB = +38dBm !!
It can be the MIC too but since the strange readings and 2,8V indicated to the magnetron I am not sure what can be exactly the problem.
The real voltage to magnetron filament, NOT under trasmission !, is 6.32Vdc
That voltage should be read from 03P9455-2 as far as I know.
Can anyone help ?
BR
******** Image 2025-12-12 at 15.04.48.jpeg******** Image 2025-12-12 at 15.04.48-2.jpeg
The peak power on the DRS6A is 5.78Kw on Long Pulse 2 (longest range). That works out to 67.62dBm, almost double your reading. What is the number on the magnetron you have in there?
Was there a resolution to this question? I'm a tech in Canada. We have what appears to be a similar situation, having replaced magnetrons (MAF1422B) on two DRS6A units on a RoRo ferry in early February 2026. Expect less than 1200 hrs. This week we get call to return. Captains are having weak and missing returns and have gain set at 100% for reasonable target recognition - on both radars.